Jump to content

Recommended Posts

Posted

Interesting case study, Ace. I'm curious how the pulse oximeter was able to give it's reading. It would appear that although this patient was in a degree of shock, they did not assess pulses manually.

I'm of the opinion there is more not being said here, than is. If anyone out there has more information regarding this, or a similar case, I would like to hear about them.

Feel free to PM, or start another thread. This is another topic that can't be discussed too often.

  • Replies 20
  • Created
  • Last Reply

Top Posters In This Topic

Posted
Interesting case study, Ace. I'm curious how the pulse oximeter was able to give it's reading. It would appear that although this patient was in a degree of shock, they did not assess pulses manually.

I'm of the opinion there is more not being said here, than is. If anyone out there has more information regarding this, or a similar case, I would like to hear about them.

Feel free to PM, or start another thread. This is another topic that can't be discussed too often.

The most likely rationale I could think of for the pulse ox still working was the patient was still utilizing compensatory mechanisims in the periphery. Another thing which wasn't mentioned was they could have placed the 'probe' centrally and gotten a reading as well... Not sure, just food for thought & my .02.

ACE844My%202c%20Large.gif

Posted

Come on now, we can't think that far ahead, remember?

The clippy thingy goes on the finger, no where else. :D

Posted
Come on now, we can't think that far ahead, remember?

The clippy thingy goes on the finger, no where else. :D

With all due respect and presesnted for your consideration, I think you should consider the following: This took place in South Africa, one of the few places on the planet where they have a doctorate program for EMS...

ACE

Posted

Whoops, ugly American there.

Must have gotten hung up on what was written, and forgot where it took place.

I would still like to hear some experiences of other providers to similar situations, though.

Posted

Pulse rate was 100 and BP 160/100. Why wouldn't the pulse oximeter work? Now a different question is would you get a reading of 100% with a resp rate of 6 and a partially obstructed airway even if you were on high flow oxygen? Only if the tidal volume was about a liter for each breath! They had problems with a consistent ETCO2 while hand ventilating which tells me they were not as careful of the rate and tidal volume of each assisted breath as they would like you to believe. I think a mechanical ventilator will give a more consistant respiratory pattern than hand bagging any time.

Just because something is published does not mean it has any validity. The follow up regarding the difference between ETCO2 and PaCO2 is important because there is a gradient between the two and it can change quickly. The gradient is dynamic which is important to remember. Unfortunately the only way to record this is to draw an ABG while noting the ETCO2 and comparing the difference. Not possible out of the hospital unless you have an iSTAT and can draw ABG's.

Live long and prosper.

Spock

Posted

For a patient being monitored by SpO2, it takes over 2 minutes of apnea before the SpO2 starts to drop. EtCO2 is almost instantaneous.

This was proven when we were being inserviced at my company for EtCO2 detectors. Since then, I rarely put my patients on SpO2.

Posted

That's correct, in fact even maybe longer (some studies describe up to 4 minutes, dependent on perfusion level). That is why I am such a proponent of EtC02 monitoring. I do believe next to ECG, this is one of the most bio monitoring devices we have. Any Paramedic that values their license, certification and does not use this or does not demand access to this is asking for problems.

What is so surprising to me, is after attending a couple of EMS conferences recently, on the number of EMS that does not offer or Paramedics that refuse to utilize this. Either this is from ignorance and poor understanding of the equipment, or plain stupidity. One thing is for sure, attorneys are beginning to understand the value of the device...... I would highly reconsider knowing all my equipment offered to me in EMS.

R/r 911

Posted

Here's another interesting use for this tool...

Hope This Helps,

2.gifACE8442.gif

(Anasthesia & Analgesia © 2005 by International Anesthesia Research Society.Optimization of Endotracheal Tube Cuff Filling by Continuous Upper Airway Carbon Dioxide Monitoring

[Technology @ Computing, and Simulation)

Efrati, Shai MD*; Leonov, Yuval MD†; Oron, Amir MD‡; Siman-Tov, Yariv DVM§,; Averbukh, Michael MD*; Lavrushevich, Alex MD||; Golik, Ahuva MD*

*Department of Medicine A, †Critical Care Unit, ‡Orthopedic Department, §Experimental Research Laboratory, and ||Department of Anesthesia, Assaf Harofeh Medical Center, Zerifin, affiliated to Sackler School of Medicine, Tel-Aviv University, Israel

Aspects of this research are disclosed and claimed in Published PCT Patent Application WO 2002/ 076279 A3. The study was supported by Hospitec Corp., the owner of this patent. SE is a shareholder in Hospitec Corp.

Accepted for publication April 4, 2005.

Address correspondence and reprint requests to Shai Efrati, MD, Department of Medicine A, Assaf Harofeh Medical Center, Zerifin, 70300, Israel. Address e-mail to efratishai@013.net.il.]

Abstract

Inappropriate cuff filling is responsible for various complications related to the use of an endotracheal tube (ETT). In this study, we evaluated an objective, noninvasive method for continuous assessment of leak around the ETT cuff by monitoring carbon dioxide pressure (Pco2) in the upper airway. Pco2 levels were measured by capnography simultaneously between the ETT cuff and the vocal cords, at the oropharynx, and in the nares of the nose. Cuff filling was regulated by an electronic controller to achieve the minimal pressure needed to prevent CO2 leak. Feasibility of the method was assessed in a human simulator and in a porcine model. Clinical function was evaluated in 60 patients undergoing surgery, comparing the method to the standard anesthesiologist evaluation. Linear correlations were observed between the ETT cuff pressure and Pco2 level in the human simulator (R2 = 0.954, P < 0.0001) and in the porcine model (R2 > 0.98, P < 0.0001). Iodine leak around the ETT cuff, in the porcine model, occurred only when Pco2 levels were >2 mm Hg. In the surgery patients, the mean ETT cuff pressure determined clinically by the anesthesiologist was significantly higher than the optimal cuff pressure assessed by Pco2 (25.2 ± 3.6 versus 18.2 ± 7.8 mm Hg, respectively; P < 0.001). According to these findings, optimal ETT cuff filling pressure can be identified by monitoring Pco2 at the nares or the oropharynx.

--------------------------------------------------------------------------------

A critical aspect in management of mechanically ventilated patients is to avoid complications related to inappropriate endotracheal tube (ETT) cuff filling (1–8). An appropriately inflated ETT cuff should achieve isolation of the lower airways, thus reducing the risk of aspiration around the cuff and possible ventilator-associated pneumonia (VAP), which occurs in up to 25% of intensive care unit (ICU) patients (5–9). However, an over-inflated cuff may cause local mechanical complications such as mucosal ulcerations, granulomas, tracheal stenosis, and tracheoesophageal fistulae (1–4).

Optimal ETT cuff filling is defined as the minimal pressure required for airway isolation. It is influenced by airway anatomy, cuff location, cuff compliance, size and volume, and by peak inspiratory pressure (10,11). The common clinical practice of optimizing cuff filling by auscultation or by assessing inhaled-exhaled volume difference is imprecise, and evaluating leak by dye infusion is impractical (12–14).

Capnographic measurement of carbon dioxide pressure (Pco2) based on infrared absorption has evolved into a commonly used procedure (15). Cyclic changes of Pco2 measured proximal to the ETT cuff can be used to identify gas leakage around the cuff. The purpose of this study was to establish an accurate, objective, noninvasive bedside method for assessment of a leak around the ETT cuff by continuous monitoring of Pco 2 at the upper airways. Initially, the feasibility of the method was investigated in a human simulator. Later, leaks at various cuff pressures were evaluated by iodine leak test in a porcine model simulating human airway mucosa. Finally, the feasibility of the method was evaluated in 60 patients undergoing elective surgery, comparing the new method to the standard clinical evaluation used today.

Methods

To find the best feasible place for CO2 measurements, we evaluated 3 anatomic locations: 1. between the ETT cuff and the vocal cords through a suction mini-guide lumen catheter attached externally to the ETT, 2. at the oropharynx above the epiglottis by a catheter inserted through a plastic oropharyngeal airway, and 3. in the nares of the nose through a nasal cannula

Figure 1. The anatomic locations of CO2 sampling in the human simulator and in patients under general anesthesia. Measurements of CO2 levels were done at 3 anatomic locations: 1. between the endotracheal tube (ETT) cuff and the vocal cords through a suction mini-guide lumen attached externally to the ETT, 2. at the oropharynx above the epiglottis by a catheter inserted into a plastic airway, and 3. at the nares of the nose through a nasal cannula.

etcapnographyleak.jpg

The study was performed in three phases. Phase 1 ascertained feasibility of the method in a human simulator. Phase 2 experimentally explored the method in a porcine model. Phase 3 compared this new method with the standard technique for estimating optimal ETT cuff filling in 60 surgical patients under general anesthesia.

Throughout all 3 phases of the study, Pco2 levels were recorded using a micro-side-stream capnograph (Microcap®; Oridion, Jerusalem, Israel) with flow rate of 50 mL/min–7.5 + 15 mL/min and rise time to CO2 step of approximately 0.2 s. By its routine use, the Microcap® uses an algorithm which detects and saves only the end-tidal Pco2 values. To avoid inaccurate processing of Pco2 data by the capnograph, the routinely used algorithm to detect end-tidal Pco2 was neutralized and all Pco2 values were directly transported and saved on a PC computer. We interpreted any increase of >1 mm Hg in Pco2 reading as a sign of leak around the cuff and accordingly set the appropriate cuff pressure. This cuff pressure was automatically kept constant by a special electronic cuff pressure controller device (TRACOE® cuff pressure controller, Germany) with a controlled accuracy of ±2 mbar, avoiding possible cuff pressure changes caused by fluctuations in inspiratory pressures or any other local variables. Throughout the experiment, we used ETTs with high-volume, low-pressure cuffs with an external side-attached mini-guide lumen which opens 1 cm above the cuff (Hi-Lo® Evac, Mallinckrodt). The proximal end of the suction mini-guide was connected to the capnograph (Fig. 1). Figure 2 shows a photograph of the PC display including a graph of the ETT cuff pressure controller and a graph of Pco2 readings in the upper airways of the human simulator.

Figure 2. Exhaled CO2 pressure waveform in accordance with endotracheal tube cuff pressures as displayed on the PC monitor in the human simulator model. Suction time (from the upper airways) is represented by the straight line of the CO2 pressure waveform.

etcapnographyleak2.jpg

Phase 1: Human Simulator Model

We used a human simulator of a 70-kg man (Medical Education Technologies, Inc. [METI], Sarasota, FL) located in the Israel Sheba Simulation Center. A Hi-Lo® Evac ETT no. 8 was inserted into the human simulator trachea. The tracheal diameter at the location of the ETT cuff was 25 mm. Exhaled air end-tidal CO2 of 40 mm Hg was simulated by adjusting a continuous CO2 stream into the simulator lungs.

While ventilating the human simulator, we sequentially inflated the cuff, 2 mm Hg at a time and measured Pco2 leak level at the 3 anatomic sites (Fig. 1). After each cuff pressure change, we suctioned the oropharynx to prevent CO2 remnants from altering subsequent measurements.

Phase 2: Porcine Model

Experiments were conducted with permission from and according to the National Institutes of Health Guidelines for the Care and Use of Laboratory Animals. Two pigs, weight 10 and 13 kg, underwent general anesthesia with a no.7 ETT. CO2 leak was assessed by a 4-mm-diameter catheter inserted directly through a small incision into the trachea 1 cm above the ETT cuff and below the vocal cords. The ETT cuff pressure was increased sequentially, 2 mm Hg at a time and Pco2 was measured through the catheter. After each measurement, we suctioned the upper airway to avoid any CO2 remnants.

After reaching a steady cuff pressure that gave no Pco2 readings at the upper airway, we surgically opened a tracheal “window” below the ETT cuff. Iodine solution (5 mL) was injected above the cuff and rested there for 60 min while we evaluated through the tracheal window any iodine leak below the cuff at different cuff pressure levels. Throughout the experiment (including the period of iodine leak measurements), animals were maintained on the same positive pressure ventilation and the ETT location and animal head and neck positions were not changed.

Phase 3: Human Subjects

Sixty consecutive adult patients undergoing elective surgery with balanced generalized anesthesia (NO2/O2) were included in our study. To obtain a homogenous group of patients regarding mechanical ventilation volumes and pressures, patients with a history of cigarette smoking or dyspnea were preassessed by pulmonary function tests. Nine patients were excluded because of forced expiratory volume per second (FEV1) or vital capacity (VC) <50% of predicted. The study protocol was approved by the local ethics committee and patients signed an informed consent.

In all patients, the tracheal intubation was performed by an anesthesiologist and the ETT position inside the trachea was determined according to the height of the patient, using the following formula: the length from the distal tip of the ETT to the right mouth angle (cm) = [body height (cm)/5] – 13 (16). After intubation, the anesthesiologist was requested to set the minimal ETT cuff pressure required to prevent leak according to exhalation-inhalation volume difference and air leak heard with the stethoscope around the cuff. The cuff pressure was considered optimal by the anesthesiologist when there was no exhalation-inhalation volume difference and no air leak heard with the stethoscope using the minimal cuff pressure for 5 min. Once cuff pressure level was considered optimal by the anesthesiologist, we started measuring Pco2 at the 3 locations: proximal to the ETT cuff via the external mini-guide lumen of the Hi-Lo® Evac ETT, at the oropharynx via a plastic oropharyngeal airway, and at the nares via a nasal cannula (Fig. 1). Based on the findings in the porcine model, optimal cuff filling was defined as the minimal cuff pressure required to avoid a Pco2 leak of >2 mm Hg proximal to the ETT cuff via the external mini-guide lumen of the Hi-Lo® Evac ETT. Suction was performed through the plastic airway as needed and 2 min before each Pco2 measurement. Continuous monitoring of Pco2 from the 3 anatomic locations was performed throughout the operation.

Statistical Analysis

The statistical SPSS software (version 10.0; SPSS Inc., Chicago, IL) was used for all analyses performed. Categorical data were expressed in numbers and percentages. Continuous data were expressed as means ± standard deviations and compared by a paired t-test. Regression coefficient ® was calculated as a measurement of correlation between ETT cuff pressure and Pco2 in the upper airways and expressed as R2. P values < 0.05 were considered significant.

Results

Phase 1: Human Simulator Model

Figure 2 illustrates an example of the exhaled Pco2 waveform in accordance with ETT cuff pressure as displayed on the PC monitor in the human simulator model. A linear correlation was observed between the ETT cuff pressure and the Pco2 measured above the cuff in all 3 anatomic locations: 1. between the cuff and the vocal cords, R2 = 0.954, P < 0.0001; 2. at the oropharynx above the epiglottis, R2 = 0.923, P < 0.0001; 3. at the nares of the nose, R2 = 0.911, P < 0.0001. Figure 3 demonstrates the correlation between ETT cuff pressures and CO2 levels measured between the cuff and the vocal cords. At an ETT cuff pressure of >>26 mm Hg, no Pco2 was recorded at any of the 3 anatomic locations (Fig. 1). At an ETT cuff pressure of 25 mm Hg, Pco2 was detected only between the cuff and the vocal cords. Once ETT cuff pressure reached 24 mm Hg, CO2 leak was measured at all locations. At all levels of cuff pressures, the maximal difference in Pco2 readings between the various anatomic locations was <2 mm Hg.

Figure 3. Correlation between endotracheal tube cuff pressures and CO2 levels, measured between the cuff and the vocal cords, in the human simulator.

etcapnographyleak3.jpg

Phase 2: Porcine Model

In animal “A,” the minimal cuff pressure needed to prevent CO2 leak around the ETT cuff was 28 mm Hg whereas the minimal cuff pressure needed to prevent iodine solution leak around the cuff was 24 mm Hg, a pressure at which we already measured a Pco2 leak of 2–3 mm Hg. There was a linear correlation between ETT cuff pressure and Pco2 leak above the cuff (R2 = 0.984, P < 0.0001)

Figure 4. Correlation between endotracheal tube cuff pressure and upper airway CO2 levels in the porcine model. The black arrow marks the point at which iodine solution leak was first detected. a, Pig A. b, Pig B.

etcapnographyleak4.jpg

In animal “B,” the minimal cuff pressure needed to prevent CO2 leak around the ETT cuff was 30 mm Hg whereas the minimal cuff pressure needed to prevent iodine solution leak around the cuff was 25 mm Hg, a pressure at which we measured a Pco2 leak of 3–4 mm Hg. There was a linear correlation between the ETT cuff pressure and Pco2 leak above the cuff (R2 = 0.988, P < 0.0001) (Fig. 4b).

Phase 3: Human Subjects

Baseline characteristics of the patients are summarized in Table 1. The mean age of the patients was 58.5 ± 16.2 yr. The mean peak inspiratory pressure was 21.2 ± 0.6 mm Hg. Although patients with severe pulmonary diseases were excluded from the study (FEV1 or VC <50% of predicted), 6 patients (10%) had mild obstructive lung disease and 1 patient had mild restrictive lung disease.

Table 1. Baseline Patient Characteristics

etcapnographyleak5.jpg

The results from the human simulator and the porcine model demonstrated a linear correlation between Pco2 leak measurements and ETT cuff pressures. The fact that iodine solution leak occurred only when the Pco2 leak readings were >2 mm Hg induced us to consider a Pco2 leak measurement clinically significant only if it was >2 mm Hg proximal to the ETT cuff via the external mini-guide lumen of the Hi-Lo® Evac ETT.

The mean initial ETT cuff pressure of all of the study population, determined clinically by the anesthesiologist, was significantly higher than the mean optimal cuff pressure determined by upper airway Pco2 leak monitoring proximal to the ETT cuff (25.2 ± 3.6 versus 18.2 ± 7.8 mm Hg, respectively; P < 0.001) (Fig. 5). In 43 patients (72%), the clinically determined ETT cuff pressure was significantly higher than the optimal cuff pressure determined by CO2 leak, with a mean change of 10.2 mm Hg in cuff pressure (25.4 ± 3.9 versus 15.2 ± 4.7 mm Hg, P < 0.0001) (Fig. 6). In 8 patients (13%) the initial ETT cuff pressures were significantly lower than the optimal cuff pressure. Nine patients (15%) had an initial ETT cuff pressure similar to the optimal cuff pressure (Fig. 6).

Figure 5. Comparison between the initial mean endotracheal tube cuff pressure, determined clinically by the anesthesiologist using the audible leak test and exhalation-inhalation volume difference, to the mean optimal cuff pressure determined by CO2 leak monitoring (n = 60).

etcapnographyleak6.jpg

Figure 6. Percentage of patients with initial endotracheal tube cuff pressure significantly higher, lower, or accurate compared with the optimal cuff pressure determined by Pco2 leak monitoring (n = 60).

etcapnographyleak7.jpg

In 3 patients, CO2 leak continued despite the exceptionally high cuff pressure, up to 35 mm Hg. The pressures determined by the anesthesiologist in these patients were 30, 32, and 33 mm Hg. After distal repositioning of the ETT, the cuff pressures needed to prevent CO2 leak were reduced to <30 mm Hg, similar to the remainder of the study population. Because cuff pressures required for complete sealing were reduced dramatically after distal reposition, we assumed that the primary incomplete sealing was probably the result of a very high proximal airway position causing poor contact between the ETT cuff and the patient airway anatomy (e.g., direct contact between the cuff and the vocal cords).

During surgery, a new leak of CO2 around the ETT cuff developed in 16 patients (27%) attributable to variable causes such as increased peak inspiratory pressure by >5 mm Hg caused by laparoscopic surgery with abdominal gas inflation, light anesthesia or inadequate neuromuscular blockade, change of head position because of surgical requirements, and ETT movement during surgery (Table 2).

Table 2. Causes of a New Leak Around the Endotracheal Tube Cuff During Operation

etcapnographyleak8.jpg

During surgery, CO2 recordings were obtained at all 3 anatomic locations (Fig. 1) and differences in Pco2 readings were <2 mm Hg. After a mean of 9 ± 4 measurements through the suction mini-guide catheter attached to the ETT, no further readings could be obtained because of obstruction by secretions and thus further Pco2 readings were obtained only from the plastic airway (hypopharynx) and from the nares.

Discussion

Methods used today to determine adequate cuff filling are either inaccurate or cumbersome (13,14). An underfilled cuff may cause aspiration of nasopharyngeal content whereas an overfilled cuff may cause local mechanical injury (1–9). The ETT bypasses normal upper airway reflexes preventing effective cough, thus facilitating pooling of nasopharyngeal secretions above the cuff which tend to “trickle” down the airway and cause VAP. Kollef et al. (17) have shown that, in patients undergoing cardiac surgery with mechanical ventilation, reducing aspiration around the ETT cuff by continuous suctioning proximal to the cuff delayed the occurrence of VAP from 2.9 ± 1.2 to 5.6 ± 2.3 days (P = 0.006). The independent risk of the ETT in ventilated patients may be deduced from studies of noninvasive ventilations (NIV). In a matched case-control study, Girou et al. (18) found that, among ICU patients, NIV reduced nosocomial pneumonia from 60% to 18% (P < 0.001). Similarly, in a large survey, 42 ICU patients receiving NIV were less likely to develop pneumonia, even after adjustment for severity of illness (19). The findings of these studies support the assumption that prevention of aspiration around the ETT cuff is crucial in order to avoid VAP.

In this study, we evaluated a new, objective, noninvasive method for determining the appropriate ETT cuff filling by continuous Pco2 monitoring at the upper airway. The method was first evaluated objectively on a human simulator model with well established anatomic structures and an unbiased environment with no secretions or nonconstant variables. The linear correlation we found between the ETT cuff pressure and CO2 leak above the cuff shows that Pco2 can be used as a quantitative indicator of air leak (Fig. 3). Although there was an overall linear correlation between ETT cuff pressure and Pco2 in the upper airways (R2 > 0.91, P < 0.0001 for all 3 anatomic locations), the correlation seems less linear at the beginning and at the end of the curve, with lower Pco2 changes than expected compared with the increase in cuff pressure (Fig. 3). This could be because, at very low cuff pressures, there is still considerable space which allows for continuous CO2 leaks, whereas, at high cuff pressures, with minimal CO2 leakage, there is a need for higher changes in cuff pressures to achieve complete sealing of the trachea by the cuff.

The human simulator represents the human airway anatomy but it cannot imitate the interactions between the natural human mucosa and the ETT cuff. Therefore, in phase 2, a porcine model was used to assess iodine leak around the cuff. Because iodine is less viscous than pharyngeal secretions, we assumed that the cuff pressure needed to prevent iodine leak would be sufficient to prevent secretion leak around the ETT cuff. In this model, using sequential increases of 2 mm Hg in cuff pressure, iodine leak around the cuff occurred only when Pco2 readings were >2 mm Hg, suggesting that the Pco2 leak test has a safety margin of 2 mm Hg. Based on these findings, and considering that maximal change in Pco2 measurements among the 3 above-cuff anatomic locations was <2 mm Hg, we concluded that clinical practice measurements of Pco2 at the oropharynx or the nares of the nose are equivalent to distal measurements just above cuff, which we found technically difficult. This second phase of the study was conducted in only two pigs because of limitations and restrictions by the National Institutes of Health Guidelines for the Care and Use of Laboratory Animals. Differences between the cuff pressure needed to prevent CO2 leak and that needed to prevent iodine leak in both pigs were almost identical, so it seemed inappropriate to kill more animals for this model.

In the third phase of the study, we evaluated the method in patients undergoing elective surgery under general anesthesia. The mean initial ETT cuff pressure determined clinically by the anesthesiologist was significantly higher than the optimal cuff pressure assessed by CO2 leak monitoring (in all patients, 25.2 ± 3.6 versus 18.2 ± 7.8 mm Hg, P < 0.001) (Figs. 5 and 6). In 72% of the patients, primary ETT cuff pressures were significantly higher than the optimal cuff pressure determined by Pco2 readings, with a mean change of 10.2 mm Hg. An over-inflated cuff increases the risk of mucosal ulcerations, granulomas, tracheal stenosis, and tracheoesophageal fistulae (1–4). The consequences of these complications are of significant clinical importance because surgical corrections are not always successful (1–4). The mechanisms are related to direct pressure necrosis by over-inflated ETT cuff duration of intubation, macro- and microtrauma during intubation, the specific technique of endotracheal intubation, severity of respiratory failure, infection, and poor tissue perfusion caused by hemodynamic instability (1–4,20). These mechanical complications are less common in elective surgical patients, and are seen mostly in patients with prolonged mechanical ventilation, such as ICU patients (1,2,4,20). The proposed Pco2 method helps determine the minimal initial cuff pressure needed to prevent a leak. However, after the initial setting, another regulatory problem, not yet solved, is the need to deflate the cuff in order to evaluate the possibility of pressure reduction (required, for example, after decreasing respiratory pressures) and then inflate it again according to the new CO2 readings. A further study is needed to evaluate the regulatory of an over-inflated cuff.

Although the general tendency of the anesthesiologists was to “overshoot” the initial cuff pressure, we found, in 13% of the cases, lower than optimal initial cuff pressures. Moreover, during surgery, a new leak developed in 27% of the patients, potentially exposing them to aspiration risk. This variability in cuff pressure, observed even in stable elective surgery patients, emphasizes the need for an accurate, continuous bedside method to determine the appropriate cuff pressure.

An important problem encountered during the study was obstruction by secretions of the mini-guide lumen of the Hi-Lo® Evac ETT after several aspirations for Pco2 measurements, despite suctioning of the upper airways. In contrast, Pco2 readings through nasal cannula, or from the oropharynx through the plastic airway, were easily obtained after secretion clearance by oral suctioning. The obstruction of the mini-guide catheter lumen, with its tip located in the trachea, by secretions, prevents its use for monitoring CO2 leak. But because the maximal difference in Pco2 readings between the 3 anatomic locations was <2 mm Hg (a pressure not causing iodine leak), we suggest the nares of the nose or the oropharynx as the locations of choice for CO2 sampling. However, we must emphasize that these findings are based on a one-size human simulator model and only 60 adult anesthetized patients. It is possible that, in patients with diverse anatomic variations, differences in Pco2 reading between the trachea and the oropharynx or the nares will be >2 mm Hg. Further study is needed to evaluate the optimal place for CO2 sampling in a larger study population with a variety of airway anatomies.

Another important issue is the position of the ETT. A malpositioned ETT is hazardous for tracheally intubated patients. Insertion of an ETT too distally leads to endobronchial intubation, which may cause collapse of the contralateral lung, whereas proximal insertion may lead to accidental extubation or vocal cord trauma (20,21). Several formulae and other methods have been proposed to estimate the optimal length for ETT insertion; however, none is always satisfactory (22–25). In the current study, 3 patients had a CO2 leak despite exceptionally high (>35 mm Hg) cuff pressures. However, pressure decreased to <30 mm Hg (as in the remainder of the study population) after distal repositioning of the ETT. Although it was not the aim of the study, continuous CO2 leak around an appropriately inflated cuff can be used as a marker for malpositioned ETT. Incessant CO2 leak around the ETT cuff can occur when there is incomplete sealing caused by poor contact with the vocal cords, endobronchial intubation (CO2 from the contralateral lung), or intubation above the vocal cords.

Regarding comparisons between the above cuff CO2 measurement and the audible leak test, several differences should be noted. Although both methods evaluate air leak around the ETT cuff, the CO2 method is an objective, accurate technique that could be automated and used as a standard in clinical trials and later as a standard of care for intubated patients, aiming to reduce complications related to the ETT tube. However, the CO2 method has several important limitations: 1. the technique requires adequate equipment, including a second capnograph (besides the capnograph used for measurements of intra-ETT CO2) and a cuff pressure controller device; 2. the need for oropharynx suctioning between samples of CO2 measurements at different cuff pressures is cumbersome and prolongs the time needed to achieve optimal cuff filling. We hope that we will soon be able to supply an automatic device capable of synchronizing the CO2 measurements, the cuff pressure controller, and the suction.

The study has several limitations: 1. the study did not include ICU patients who required prolonged mechanical ventilation. 2. The experiment was performed only during positive pressure ventilation whereas the CO2 method may not be as reliable in spontaneously breathing patients. 3. Patients with moderate-to-severe pulmonary disease were not included in the study. However, there is a real need for a simple, noninvasive method that can be used as an objective standard in clinical practice and in studies addressing the important issue of leak around the cuff. This standard may be applied not only to short-term over-inflation of ETT cuffs in the operating room, which has less significant clinical importance, but also to more complicated conditions, such as in neonates, children, patients with pulmonary diseases or anatomic malformations, and ICU patients. However, before using this new method in more complicated conditions, we evaluated the method in optimal conditions.

In this study, we evaluated a new, objective method for optimizing ETT cuff filling by monitoring CO2 levels in the upper airways. The method is simple, noninvasive, and can be used in any tracheally intubated patient. Continuous CO2 monitoring in the upper airway can be used to identify the minimal cuff pressure needed to prevent an air leak around the ETT cuff. Data obtained in two pigs suggest that this method may also be useful for identifying the cuff pressure that would prevent aspiration of secretions around the cuff. Before the CO2-based cuff inflation method can be use as a standard of care for all tracheally intubated patients, further studies comparing the clinical relevance of the differences between the audible leak test to the CO2 method are needed for patients more prone to ETT complications, such as those with pulmonary diseases, ICU patients, and children.

References

1.Stauffer JL, Olson DE, Petty TL. Complications and consequences of endotracheal intubation and tracheotomy: a prospective study of 150 critically ill adult patients. Am J Med 1981;70:65–76. SFX Bibliographic Links Library Holdings [Context Link]

2.Lewis FR Jr, Schiobohm RM, Thomas AN. Prevention of complications from prolonged tracheal intubation. Am J Surg 1978;135:452–7. SFX Bibliographic Links Library Holdings [Context Link]

3.Klainer AS, Turndorf H, Wu WH, et al. Surface alterations due to endotracheal intubation. Am J Med 1975;58:674–83. SFX Bibliographic Links Library Holdings [Context Link]

4.Ferdinande P, Kim DO. Prevention of postintubation laryngotracheal stenosis. Acta Otorhinolaryngol Belg 1995;49:341–6. SFX Bibliographic Links Library Holdings [Context Link]

5.Scheld WM. Developments in the pathogenesis, diagnosis and treatment of nosocomial pneumonia. Surg Gynecol Obstet 1991;172:42–53. SFX Bibliographic Links Library Holdings [Context Link]

6.Garrouste-Orgeas M, Chevret S, Arlet G, et al. Oropharyngeal or gastric colonization and nosocomial pneumonia in adult intensive care unit patients. Am J Respir Crit Care Med 1997;156:1647–55. SFX Bibliographic Links Library Holdings [Context Link]

7.Guidelines for prevention of nosocomial pneumonia. Centers for Disease Control and Prevention. Respir Care 1994;39:1191–236. SFX Library Holdings [Context Link]

8.Spray SB, Zuidema GD, Cameron JL. Aspiration pneumonia: incidence of aspiration with endotracheal tubes. Am J Surg 1976;131:701–3. [Context Link]

9.Craven DE, Steger KA. Epidemiology of nosocomial pneumonia: new perspectives on an old disease. Chest 1995;108:1–16. [Context Link]

10.Kearl RA, Hooper RG. Massive airway leaks: an analysis of the role of endotracheal tubes. Crit Care Med 1993;21:518–21. SFX Bibliographic Links Library Holdings [Context Link]

11.Guyton DC, Barlow MR, Besselievre TR. Influence of airway pressure on minimum occlusive endotracheal tube cuff pressure. Crit Care Med 1997;25:91–4. Ovid Full Text Bibliographic Links Library Holdings [Context Link]

12.Fisher MM, Raper RF. The “cuff-leak” test for extubation. Anaesthesia 1992;47:10–2. SFX Bibliographic Links Library Holdings [Context Link]

13.Petring OU, Adelhoj B, Jensen BN, et al. Prevention of silent aspiration due to leaks around cuffs of endotracheal tubes. Anesth Analg 1986;65:777–80. SFX Bibliographic Links Library Holdings [Context Link]

14.Young PJ, Basson C, Hamilton D, Ridley SA. Prevention of tracheal aspiration using the pressure-limited tracheal tube cuff. Anaesthesia 1999;54:559–63. SFX Bibliographic Links Library Holdings [Context Link]

15.Schmitz BD, Shapiro BA. Capnography. Respir Care Clin North Am 1995;1:107–17. [Context Link]

16.Cherng CH, Wong CS, Hsu CH, Ho ST. Airway length in adults: estimation of the optimal endotracheal tube length for orotracheal intubation. J Clin Anesth 2002;14:271–4. SFX Bibliographic Links Library Holdings [Context Link]

17.Kollef MH, Skubas NJ, Sundt TM. A randomized clinical trial of continuous aspiration of subglottic secretions in cardiac surgery patients. Chest 1999;116:1339–46. Ovid Full Text Bibliographic Links Library Holdings [Context Link]

18.Girou E, Schortgen F, Delclaux C, et al. Association of noninvasive ventilation with nosocomial infections and survival in critically ill patients. JAMA 2000;284:2361–7. Ovid Full Text Bibliographic Links Library Holdings [Context Link]

19.Carlucci A, Richard JC, Wysocki M, et al. Noninvasive versus conventional mechanical ventilation: an epidemiologic survey. Am J Respir Crit Care Med 2001;163:874–80. SFX Bibliographic Links Library Holdings [Context Link]

20.Streitz JM Jr, Shapshay SM. Airway injury after tracheotomy and endotracheal intubation. Surg Clin North Am 1991;71:1211–30. SFX Bibliographic Links Library Holdings [Context Link]

21.Cavo JW Jr. True vocal cord paralysis following intubation. Laryngoscope 1985;95:1352–9. SFX Bibliographic Links Library Holdings [Context Link]

22.Owen RL, Cheney FW. Endobronchial intubation: a preventable complication. Anesthesiology 1987;67:255–7. SFX Bibliographic Links Library Holdings [Context Link]

23.Mehta S. Intubation guide marks for correct tube placement: a clinical study. Anaesthesia 199;46:306–8. [Context Link]

24.Patel N, Mahajan RP, Ellis FR. Estimation of the correct length of tracheal tubes in adults. Anaesthesia 1993;48:74–5. SFX Bibliographic Links Library Holdings [Context Link]

25.Cherng CH, Wong CS, Hsu CH, Ho ST. Airway length in adults: estimation of the optimal endotracheal tube length for orotracheal intubation. J Clin Anesth 2002;14:271–4. SFX Bibliographic Links Library Holdings [Context Link]

Is Capnometry the Optimum Method for Assessing the Adequacy of Endotracheal Tube Cuff Seal?

[Letters to the Editor]

Sosis, Mitchel B. MS, MD, PhD

Board Certified Anesthesiologist, Lafayette Hill, PA, mitchelsosis@hotmail.com

To the Editor:

Efrati et al. (1) recently proposed a new technique for assessing the adequacy of endotracheal tube cuff seal on the basis of upper airway capnometry at three sites. The cuffs were inflated until a leak of <=2 mm Hg of carbon dioxide was detected. This was compared to the “standard” technique of filling the cuffs by listening for a leak and by comparing inspired and expired tidal volumes. They concluded that capnographic sampling of carbon dioxide via a nasal cannula could “objectively” detect an inadequately inflated endotracheal tube cuff in a patient undergoing general orotracheal anesthesia.

Carbon dioxide is not always detected via a nasal cannula in spontaneously breathing non-intubated patients. The reason, of course, is that the patient can choose to breathe through the mouth. Unless the mouth is somehow sealed in intubated patients, it seems that there is no guarantee that a nasal cannula would detect carbon dioxide leaking past the endotracheal tube cuff.

In the clinical limb of this study, Efrati et al. compared the proposed capnometric technique to the standard technique. The authors noted a slightly lower mean cuff pressure (25.2 ± 3.6 versus 18.2 ± 7.8) with capnography. However, it is not clear whether clinicians were blinded as to cuff pressures, and/or attempted to minimize those pressures, when using the standard technique. The authors did not assess postoperative sore throat, aspiration, or other sequelae of endotracheal intubation. Consequently, no clinical benefit of the capnometric technique was demonstrated.

Mitchel B. Sosis, MS, MD, PhD

Board Certified Anesthesiologist

Lafayette Hill, PA

mitchelsosis@hotmail.com

Reference

1. Efrati S, Leonov Y, Oron A, et al. Optimization of endotracheal tube cuff filling by continuous upper airway carbon dioxide monitoring. Anesth Analg 2005; 101:1081–8. Ovid Full Text Bibliographic Links Library Holdings [Context Link]

Posted

OK now time out. I agree with the importance of ETCO2 monitoring but it does not replace pulse oximetry. You need both pieces of information to assess your patient. ETCO2 measures ventilation while pulse oximetry measures oxygenation which are two different things. Pulse oximetry will drop almost immediately in patients with obstructive or restrictive lung disease or those with reduced functional residual capacity. It will stay at 100% for a young and healthy patient for a long time before dropping.

The best way to tell if your patient is apneic is to look at them. The best monitor is a vigilant health care provider.

Live long and prosper.

Spock

This thread is quite old. Please consider starting a new thread rather than reviving this one.

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.


×
×
  • Create New...